The Center for Devices and Radiological Health (CDRH) has adopted the USP Endotoxin Reference Standard and limits for medical device extracts expressed in EU/mL. USP Chapter <161> Transfusion and Infusion Assemblies and Similar Medical Devices provides the limits for medical devices within its scope. The endotoxins limit for a medical device is dependent on the intended use of the device and what the device contacts (e.g., blood, the cardiovascular system, cerebrospinal fluid, intrathecal routes of administration, permanently implanted devices, and devices implanted subcutaneously).
For medical devices, using the extraction volume recommendations described below, the limit is 0.5 EU/mL or 20 EU/device for products that directly or indirectly contact the cardiovascular system and lymphatic system. For devices in contact with cerebrospinal fluid, the limit is 0.06 EU/mL or 2.15 EU/device. For devices that are in direct or indirect contact with the intraocular environment, a lower endotoxins limit may apply. Please contact the appropriate review division for specific recommendations.
The process of preparing an eluate/extract for testing may vary from device to device. Some medical devices can be flushed, some may have to be immersed, while others may need disassembly. Unless otherwise directed by another compendial standard, our recommended rinse volumes include the following: (1) each of the 10 test units should be rinsed with 40 mL of non-pyrogenic water; (2) for unusually small or large devices, the surface area of the device that contacts the patient may be used as an adjustment factor in selecting the rinse or extract volume. The endotoxins limit can be adjusted accordingly. In any case, the rinse/extract procedure should not result in a greater dilution of endotoxin than recommended in USP <85>. For inhibition/enhancement testing, both the rinse/extract solution and the device eluate/extract should be tested.
Examples of medical devices with testing or interference challenges include devices that are coated with anticoagulant, contain heavy metals, or that have particulates. In these situations, treatments for interferences can include digestion, dilution, and addition of buffers, centrifugation, or filtration.
During the same surgical procedure or placement in the same surgical site, multiple units of the same device from one manufacturer should generally meet the same endotoxins limit as a single device administered during the procedure. In instances where multiple units of the same device are known or intended for use in a single procedure, manufacturers should justify any deviation from the overall endotoxins limit identified in this guidance.
When a manufacturer of medical devices plans to use LAL testing that deviates significantly from this guidance or recognized standard, a premarket notification (510(k)) under section 510(k) of the Federal Food, Drug, and Cosmetic Act (the Act) or a premarket approval application (PMA) supplement under section 515 of the Act should be submitted. Significant deviations include, but are not necessarily limited to: higher endotoxin concentration release criteria, sampling from fewer than three (3) lots for inhibition/enhancement testing, lesser sensitivity to endotoxins, and a device rinsing protocol resulting in greater dilution of endotoxins than that recommended in this guidance.